Post by jianmei on May 14, 2018 23:32:18 GMT -5
Hi, I have a L335, and when I tried to increase the temperature from 10K to 450K, the heater automatically turned off. The heater can only keep open for about 7 seconds. And after it's closed, the interface shows "Heater 1 Open". But the heater doesn't work. Can you tell me how to fix it? Thanks!
Jianmei
|
|
|
Post by Lake Shore Jeff M on May 15, 2018 7:48:00 GMT -5
Jianmei,
The Heater 1 Open error is saying that the Model 335 has detected that the resistance of the heater exceeds approximately 500 ohms. When you start the heater, do you see the temperature of the control sensor start to increase and then the error appears? If yes, what may be happening is that as the system starts to warm up, one of the wires going to the heater is going open which causes the error. If you do not see the temperature start to go up before seeing the error it may indicate an issue in the Model 335 heater circuit.
The first thing to do is to disconnect the heater from the rear of the Model 335 and measure the resistance across the prongs of the dual banana plug looking back at the heater. If it reads greater than 500 ohms, you will need to investigate that circuit. If it reads normal values which are normally around 25 to 50 ohms, reconnect it and try to do an experiment. Monitor the temperature and see if the open error appears at he same temperature each time. If it does, then the open is due to an issue in the system you are using and you will need to contact that manufacturer for troubleshooting procedures. If not, let me know the details of the failure and we can do additional Model 335 troubleshooting.
|
|
|
Post by jianmei on May 15, 2018 18:31:25 GMT -5
Jeff, Thank you. The problem has been solved.
|
|
ryann
New Member
Posts: 1
|
Post by ryann on Nov 24, 2022 9:07:33 GMT -5
Hi jianmei I have the same problem, how do you solve it ?
|
|
|
Post by shmah1 on Mar 27, 2023 9:51:53 GMT -5
Hi Jeff M, I am going through the same problem. It is reading 25 ohm and still showing the error heater 1 open. I reconnected it several times. Can you please help me in this case?
|
|
|
Post by Lake Shore Ogi on Mar 27, 2023 10:54:58 GMT -5
Dear shmah1:
There are several reasons that will trip this error. Please check for the following and send me an email to support@lakeshore.com with your findings:
1. Model and serial number of your controller 2. Do you have a valid sensor reading for the sensor that is configured for the control loop (or is the sensor showing an error such as S.OVER, T.OVER)? 3. Does this error occur in all or just specific heater output ranges? 4. Does your error pop up immediately after you turn the heater on or does it take some time to trip?
Thank you, Ogi K.
|
|