Post by ivanhoe on Jun 2, 2023 11:13:40 GMT -5
Hello I have problem with LS330 it stops to put power for heater. When I set any power range (low, med or high) I can hear the characteristic "click" inside the controller but no output power, the front indicator shows nothing and no voltage between "high" and "low" on the rear panel. (The resistance of the heater is 25 Ohm) . Is there anything I can do? Thank you
|
|
|
Post by Lake Shore Jeff M on Jun 2, 2023 11:54:30 GMT -5
Providing you have a P value that is greater than 1 and your setpoint is higher than the control sensor temperature, if you have no output, there is most likely a defective component in the heater circuit and there nothing you can do other than purchase a new controller as we no longer offer repair service for the Model 330.
I suggest you reset the controller to factory default values, then reconfigure the controller and if it still does not work, there is nothing you can do other than replace the controller.
To reset the controller, press and hold the Escape and Units buttons in for approximately 20 seconds and when all of the LED's light release the buttons.
|
|
|
Post by ivanhoe on Jun 2, 2023 15:01:03 GMT -5
Hello Jeff Thank you very much for your response! I will try your "recipe"
|
|
|
Post by ivanhoe on Jun 2, 2023 15:19:11 GMT -5
Thank you Jeff It is working now! We were working with this device remotely (GPIB LabView) and accidentally P value was set to 0. Have a great weekend!
|
|