Post by ascender on Aug 29, 2017 4:35:24 GMT -5
Dear Sir,
We are client from Hong Kong. Our machine was purchased and received on May 2017
After sent the gauss meter 425 to HUA NAN CALIBRATION LIMITED for calibration. The accuracy was unacceptable. It is about 3-4% error.
When using the 2504-VR probe, the reading of zero seem not stable and keep have some reading.
We have further verify the result by using MRA-312-2K (2kG magnet) and found the value is 3 % less than actual.
When replace the probe with our old 2502-VR probe (Machine stated incompatible but still can have reading) the reading seem correct and the zero was steady.
Pls advise is there any setting missing. (Already set zero and have try for setting the probe field on and off )
Pls advise how to sent back whole set of gauss meter and probe for repair or replacement. (sent to Shang Hai agent or US Lakeshore?)
Will the freight charged by Lakeshore?
|
|
|
Post by Lake Shore Jeff M on Aug 29, 2017 6:19:32 GMT -5
Greetings Ascender,
LinkPhysics Corporation, our representative for Chine reported this issue to our service department yesterday. As I informed them, there are no settings that can be used to adjust the probe and based on the fact that the old probe and the Model 425 Gaussmeter reads the field correctly, I recommended that the Hall Probe be returned for a Warranty Failure Inspection. As long as the probe has not been damaged, we will repair and recalibrate or replace the probe and return it to you under warranty.
As defined in our warranty policy, you are responsible for the shipment to Lake Shore. As discussed above, providing no damage is found, we will repair, recalibrate and/or replace the probe at no additional cost to you.
Please continue to work with LinkPhysics to arrange the return for inspection.
|
|