I calibrated a Cernox sensor and entered the curve into my instrument, however, I do not see it as a curve selection when configuring the input. Why is this?
Each sensor type requires a specific curve format. In the case of a Cernox® sensor, the curve format must be in LogΩ (Base 10) vs. K. If you entered the curve in the resistance vs. temperature format, it will not appear as a selection.
|
|