Post by yhbyhb4433 on Nov 29, 2018 17:37:11 GMT -5
Hello, my Cernox is malfunctioning, and I'm trying to figure out what could have caused the problem. I'm seeing a current of 15-20uA at 295K (which is misleading because it's supposed to be 120ohm at 300K for 2mV excitation). But at 78K, it still has a current of 10uA. This happens after I soldered the leads onto PCB pads. I don't think the solder iron was applied for more than a couple of seconds, but it seems the Cernox is malfunctioning since. Are there any failure mode that could cause Cernox to have similar current across temperatures? Thank you.
|
|
|
Post by Lake Shore Jeff M on Nov 30, 2018 8:12:57 GMT -5
In order to assist in the troubleshooting, I will need additional information. Please let me know the following: Model and/or serial number of the sensor How you are making the leakage current measurements What is the resistance between leads I+ & V+, I- & V-, V+ to V- Also, measure between each lead and the sensor package or the chassis if the sensor is mounted.
With this information, I can determine what may be causing the leakage.
|
|