Post by suplab on Dec 19, 2019 12:12:56 GMT -5
Hi
1. I needed to know is there any issue if we got the +ve and and -ve terminals reversely connected to the temperature controller for current or voltage leads. (Note: The calibration was done with the right leads connected to right terminals.)
2. Again is there any issue if the CX-1050-AA sensor used in vacuum environment.(Note: The calibration was done in He environment.)
Thank you
|
|
|
Post by Lake Shore Jeff M on Dec 19, 2019 12:34:57 GMT -5
The sensor is not polarity sensitive, however, the measurement equipment will be, therefore, you must have both positive leads on one side of the sensor and both negative leads on the other side in order to properly interpolate the temperature.
The "AA" package is vacuum compatible, therefore, there should be no issue providing it is properly installed.
|
|