Post by hamzastro on Aug 12, 2020 6:15:51 GMT -5
Hi, I have a Pt lead on a Cernox sensor (CX-1050-SD-HT-1.4L-P) which has been detached from the sensor because I needed to bend it for mounting purpose. Is it possible to repair it in the lab or should I send it for repairing ?
If possible in the lab, can you please suggest a procedure for that ? (time and energy of spot-welding, and any other recommendations). I hope this is not the end of the sensor !
Thank you,
|
|
|
Post by Lake Shore Jeff M on Aug 21, 2020 9:36:29 GMT -5
The following information was provided via email with the disclaimer, "We are ultimately not responsible for what happens if the attempted repair is unsuccessful"
Suggestions: • Do not remove the residual piece of Pt wire on the bond pads, but look to the side of it where there is open pad space. o If this is really beat up, do not attempt repair. Order replacement. o If there is good bondable metallization on the remaining bond pad, then that is where you want to make a new weld. • If re-weld is doable, straighten out the ribbon wires and clean with high purity Isopropyl alcohol. Dry with Kimwipe/labwipe and allow to air dry for a while. Clean the bond pads with high purity Isopropyl alcohol and acid brush. Dry in the same manner. o Wear gloves and don’t touch parts with hands. • Use a DC welder, something like a Unitek PECO ThinLine, with a parallel gap type tip. Select the correct tip material for precious metals. • Dress the tip using a very flat piece of Alumina. There should be steps on web to show how to do this. o Make sure not to touch the tip with bare hands, wear gloves. Check the tip for any FOD and remove before welding. • Set the DCV pulse amplitude, duration (usually millisec), and downward force to apply with tip. • Align ribbon wire in new bonding location and bring tip down to create a new weld. o BTW, this is something that the operator usually has to build a weld schedule for. So this could take several tries. It would be much better if the customer had similar trial/scrap materials to help dial-in the weld schedule needed for proper repair.”
|
|