Post by thomas on Oct 15, 2020 12:36:07 GMT -5
I am currently having an issue where when I attempt to power on our Model 340 Temperature controller by turning on the switch on the back panel, the unit does not show any signs of turning on and immediately blows the fuse next to the power switch. I have replaced the fuse with the exact 120 V 2 A slow-blow fuse described in the manual multiple times but every time I try to turn it on I get no response and a blown fuse.
Do you have any idea what could be causing this issue that the unit will not even turn on without blowing a fuse? I have disconnected every input and output to try and isolate the issue to startup alone.
I would appreciate any thoughts or suggestions you might have to resolve this.
Thanks and best regards,
Thomas
|
|
|
Post by Lake Shore Jeff M on Oct 15, 2020 13:31:32 GMT -5
Thomas, basically the model 340 has developed an issue in the power circuit of the main PCB and has most likely caused the power transformer to overheat and short out the input windings causing the fuse to blow.
Depending on the age of your Model 340, will depend on what needs to be done. I suggest you consider sending the Model 340 back to Lake Shore for a repair which will cost $1683.00 plus shipping and normally is completed in about 2 weeks. However, I can tell you that depending on the failure, it is possible that a repair will not be successful so you may want to consider purchasing a new controller instead of spending $1683.00
If you want to return the controller for a repair, please contact us at support@lakeshore.com and provide the serial number of the Model 340 as well as your complete contact information. We will send you a Return Authorization with additional instructions.
|
|