Post by sarendt on Oct 26, 2020 17:50:17 GMT -5
I am communicating with a Lakeshore 350 from a LabVIEW app. Among other things, I am using the HTR command to get the heater output. The response to the HTR command gives the heater output with a precision of 0.1 whereas the instrument front panel gives the heater output with a precision of 0.01. Is there a way to get the heater output with full precision (0.01) using the HTR command (or some other command)?
|
|
|
Post by Lake Shore Jeff M on Oct 27, 2020 8:38:01 GMT -5
The resolution over the interface is limited to 6 characters which includes the polarity, 3 digits for the value, 1 digit for the decimal point and the last one for the decimal value. There are no other commands that would give you more resolution via a command prompt.
|
|
|
Post by sarendt on Oct 27, 2020 9:36:34 GMT -5
OK, thx for the reply
|
|