Post by jf on Dec 29, 2021 1:41:45 GMT -5
Hi,
We have a Lakeshore CRX-VF connected to the model 336 temperature controller. However, recently, the temperature sensor for input A showed a T.OVER reading. We read that this happens when the temperature goes beyond the range possible measured by the sensor. However, the stage is currently set at room temperature so there must be something wrong with the sensor setup.
Is there a way to troubleshoot and identify which part of the setup is causing the issue? We are using diode based temperature sensor GaAlAs which I think is one of the sensors that Lakeshore provides.
Thank you.
|
|
|
Post by Lake Shore Jeff M on Jan 5, 2022 14:44:23 GMT -5
Based on the information provided I suspect that the sensor has failed or the wiring has come loose causing the sensor leads to short. This issue has been transferred to our customer support team as case # 13401.
|
|