Post by painter on Mar 21, 2022 12:19:10 GMT -5
We are mounting a cernox sensor with the CU package onto an electrical bus. Do we need to electrically insulate the Cu package from the bus or is the cernox already electrically insulated from the CU package?
Thanks,
- Tom
|
|
|
Post by Lake Shore Jeff M on Mar 21, 2022 16:24:50 GMT -5
The sensor inside the CU Package is insulated from the package. You can verify this by measuring the resistance between each of the leads to the copper bobbin which should be open.
|
|