Post by s3 on Jan 4, 2023 9:51:21 GMT -5
Bonjour, je voudrais savoir si la sonde cernox XCX-1050-BO-1.4L-145591 peut être interfacée avec le régulateur de température 218? Je viens de fixer la sonde, et je ne peux pas l’interfacer. Je voudrais être sûr qu’elle est bien connecté, je mesure actuellement 10ohms à température ambiant Merci pour vos commentaires S3
|
|
|
Post by Lake Shore Jeff M on Jan 4, 2023 11:28:00 GMT -5
Yes, the XCX-1050-BO-1.4L sensor can be interfaced with the Model 218 monitor. However, your room temperature reading of 10 ohms sounds to be low.
You will need to use the Curve Handler application to load the curve for your sensor into the monitor into the curve location for the Input you are connecting the sensor to. Input 1 uses location 21 up to Input 8 which uses location 28.
The sensor connection requires a 4-wire connection as seen in section 3.3.2.5 of the Model 218 User Manual. The lead assignment to the 25-pin connector is shown in Figure 3 of the manual.
If you tell me what the serial number is for the sensor you have I can tell you what resistance to expect at room temperature.
I have attached our RTD sensor Troubleshooting Guide so you can verify the sensor is wired correctly. You can send the results to support@lakeshore.com once you have all of the readings in place.
|
|
s3
New Member
Posts: 2
|
Post by s3 on Jan 5, 2023 5:03:04 GMT -5
Thank you for your feedback, I finally have a consistent measurement with the 218 controller Sincerely
|
|