Post by jacob on Jan 11, 2023 20:09:00 GMT -5
Hi, hope you're having a good day!
I am using the Cryotronics 8400 Series Hall measurement system, and am hoping to calibrate / confirm the accuracy of the measured data.
So far I have used to probes with the 0.1 ± 1% Ohm and 1 ± 0.05% Ohm resistors, and used IV measurements to confirm the accuracy of the measured voltage and current.
I noticed the 'FIELD CALIBRATION CARD' and am curious if I may use this to calibrate / confirm the accuracy of the Hall Probe measuring the magnetic field. If so, how may I do so?
Thank you, Jacob
|
|
|
Post by LakeShore Chris on Jan 24, 2023 14:11:03 GMT -5
Hello Jacob,
That is exactly what the calibration card is for. You will insert the card into the r/t sample insert, from the 8400 software you will go to the toolbox tab and select Field calibration. On the cal card there should be a calibration coefficient that you will enter into sensor sensitivity box that will come up, from here select the cal you want to perform.
|
|