Post by emilio on Feb 8, 2023 9:50:12 GMT -5
Hello,
We have quite some lakeshore modules in our installation but when connecting the grounding pin to the cabinet ground bar the resistance increases from 0.1 to 0.5 Ohm. Can you please explain what can be the cause?
Thank you.
|
|
|
Post by emilio on Feb 8, 2023 9:52:33 GMT -5
|
|
|
Post by Lake Shore Jeff M on Feb 8, 2023 10:12:55 GMT -5
The Model 240 acquires its ground through the DIN Rail connection. The ground connection on the Input is for the shield of the cable and should not be connected to the cabinet grounding bar as that will create a ground loop.
|
|
|
Post by Lake Shore Jeff M on Feb 8, 2023 10:13:16 GMT -5
The Model 240 acquires its ground through the DIN Rail connection. The ground connection on the Input is for the shield of the cable and should not be connected to the cabinet grounding bar as that will create a ground loop.
|
|
|
Post by emilio on Feb 9, 2023 7:26:26 GMT -5
Hi Jeff, we tried that at the beginning but without the shield pin connected to the cabinet grounding the signal was very noisy. Have you experienced that before?
|
|
|
Post by Lake Shore Jeff M on Feb 9, 2023 14:03:33 GMT -5
The only time I have experienced this is if the rack is not grounded which means the DIN Rail is not grounded or the cable shield wire was grounded at both ends. Make sure that the Ground connector is present on the 240 which is a silver clip on the DIN Rail mounting location opposite of the spring loaded clip.
|
|
|
Post by emilio on Feb 10, 2023 8:59:14 GMT -5
Thanks Jeff! Will talk with the team and come back if we have more questions, highly appreciated.
|
|
|
Post by jhonmartin on Sept 16, 2023 13:30:46 GMT -5
The increase in resistance when connecting the grounding pin to the cabinet ground bar might be due to various factors. It could be caused by loose connections, corrosion, or even the quality of the ground bar itself. I recommend checking and ensuring that all connections are secure and clean 3d result. If the issue persists, it might be a good idea to consult with a qualified electrician or technician to diagnose the specific cause and address it effectively.
|
|
|
Post by storiesigapp on Jun 14, 2024 4:07:20 GMT -5
The increase in resistance when connecting the grounding pin to the cabinet ground bar from 0.1 to 0.5 Ohm could be caused by several factors: Contact Resistance: Poor contact between the grounding pin and the cabinet ground bar can introduce additional resistance. Ensure the connections are clean, tight, and free of corrosion. Ground Loop: Connecting multiple grounding points can create a ground loop, leading to increased resistance. Check this
|
|