Hi: Hope you have a nice day. I am a new learner for flux testing, for FP series hall probe, questions as below, hope you can give me answer. 1. for FP series hall probe, which kind of signal output, voltage or current? 2. I want to connect the FP series hall probe with my existed amplifier, such NI and HBM, but what is the Hall probe connector pins definition(not find in this website)?
|
|
|
Post by Lake Shore Ogi on Jun 16, 2023 14:29:34 GMT -5
Dear Customer,
Thank you for reaching out to us through our forum! I can assist you with general information regarding the pin-outs and current drive. However, I want to highlight that using a traditional current source and voltage input to obtain usable measurements from the FP probe may not be feasible. The F41/F71 utilizes a current-spinning technique to eliminate offset, and I anticipate that the offset you would measure with your setup would be significantly large. Additionally, without a calibration reference, the probe measurements would lack meaningful interpretation.
If you still wish to proceed with connecting the probe to your equipment, kindly send me the model number of your probe via email at support@lakeshore.com. I will then provide you with the appropriate pin-out details for the probe series. The current drive for the probe is 1mA.
Greetings, Ogi K.
|
|