Post by labmaster on Jul 2, 2023 0:23:32 GMT -5
Hi,
I have a temperature sensor but two outputs in 335. (Actually another sensor was out of order now.)
I found I cannot share the sensor with two 4-lines for A/B input channels or it seemed not to be set with one shared input for A/B outputs in internal setting.
Is it right or suggest to other way for two (same) outputs with one sensor?
Best,
labmaster.
|
|
|
Post by Lake Shore Jeff M on Jul 2, 2023 21:03:52 GMT -5
What you have found is correct, an input can only be assigned as the control input for one of the outputs at any given time.
|
|
|
Post by Lake Shore Ryan on Jul 5, 2023 10:44:56 GMT -5
Hi labmaster, do you have two separate heaters providing heat to the same location where the sensor is installed?
|
|