Post by nikola on Feb 21, 2024 13:29:50 GMT -5
Hi,
I use temperature controller 350. On channel A I have Pt100 and channel B NTC100kOhm. The problem is that on channel B after exceeding the temperature limit of channel B, the resistance of the NTC is measured wrong. I measured via a multimeter and Keithley 2002, both show resistance of 112kOhm at room temperature. However, LakeShore measures 3kOhm??? If I switch the channels to and measure Pt100 on channel B and NTC100kOhm on channel A is the same channel A shows wrong values for NTC while channel B is correct for Pt100.
I tested on another controller also a 350 model and have the same issue (both can not be damaged)!? But on 334 model it shows the correct resistance for NTC100kOhm. Any idea, how to manage the troubleshooting?
|
|
|
Post by Lake Shore Jeff M on Mar 8, 2024 10:25:41 GMT -5
The Model 350 needs to be configured for a 10mV excitation to read MTCRTD resistance above 100K ohms. Were you using the 1mV excitation range when reading the sensor when you saw the incorrect value?
|
|