Post by jurumy7 on Aug 20, 2024 21:10:18 GMT -5
Hi everyone, For what I understand, the shield pin (pin 3,6 of DIN connector) should be isolated from the chassis ground (=power cord ground prong). But when I measure my 336 controller (with all the inputs and power cord, connection cable detached), it shows 1 Megaohm between shield pin and chassis ground, not Over limit. Why is this happening? Wouldn't this interfere my cryostat's ground as my cryostat's ground (=connected to clean ground which is isolated from power outlet grounds) would be connected to chassis ground by the cable shield, as the 336 manual says one shield should be connected to the cryostat's ground?
I am using 218 & 331 & 336 together for my measurement configuration, so I need to understand why this is happening and how the 336 is different from the others.
Thank you in advance.
|
|
|
Post by Lake Shore Jeff M on Aug 21, 2024 7:08:02 GMT -5
By design, the measurement ground in the Model 336 is isolated from the chassis using a 1 Meg Ohm resistor in parallel with an 0.1µf capacitor. The shield pins tie to the measurement ground which is why you are reading the 1 Meg Ohm value between the pins and the instrument ground.
The Model 331 has a similar design and the Model 218 does not have any isolation between the measurement ground and the chassis.
In theory, the 1 Meg Ohm isolation is sufficient to prevent ground loop issues.
|
|
|
Post by jurumy7 on Aug 21, 2024 21:08:43 GMT -5
Thank you Jeff!!
|
|