I'm looking to connect a Cernox RTD (SD package) within a cryostat that has an electrical feed through with a D type connector. I have 32AWG Quad-Twist lead wire that I'm looking to bridge the gap between RTD and D connector inside the cryostat. Is 32AWG wire too small for the solder cups? Is this a typical arrangement?
Thank you.
|
|
|
Post by Lake Shore Jeff M on Feb 27, 2018 7:38:25 GMT -5
Jeff,
The 32AWG Quad Twist Phosphor Bronze wire is suitable to be soldered to the connector you are using. The soldering technique will be similar to attaching the leads to the sensor which is described in our Cernox™ Sensor Installation Instructions at:
www.lakeshore.com/Documents/F017-00-00.pdf
Let me know if you have additional questions.
|
|